On 4/3/2015 8:40 PM, Josh Dersch wrote:
So I may have been a bit premature in my declaration
earlier this week
that letting the 4014 warm up for a few minutes solves the storage
problems.
For the time being, I have it hooked up to a Linux box (so I can 'cat'
various files at it and stare in awe as it draws random things) and it
seems to be performing flawlessly; everything works (including the
discrete plot extensions). But I've noticed that over time as I clear
the screen that garbage starts accruing around the edges of the screen
-- only the middle gets properly erased.
At first power-on, the area that gets cleared is a circle maybe 10" in
diameter; this increases slowly over time and if I let it run for
15-20 minutes *most* of the screen gets cleared but there's always a
bit on the edges that remains.
I went through the portions of the alignment procedure outlined in the
service manual related to storage, and all voltages were within a
percentage point or two even after all these years, so not much
required adjustment.
There are two adjustments for the Collimation that control the size
and shape of the flood that erases the screen; the service manual
suggests adjusting these until the flood covers the screen. Adjusting
the pots all the way counter-clockwise causes the flood to get
*slightly* larger and cover more of the screen, but it's still not
enough.
From reading the circuit description for the erase circuits (starting
on page 5-82 of the service manual), I note that the duration of the
flood (as controlled by the Collimation circuits) is controlled by an
RC network and I suppose it's possible that one or more of the
capacitors is out of spec -- but I don't know if the length of the
flood has anything at all to do with the area it covers -- can anyone
shed some light on this?
I suppose it's more likely that the tube's just showing its age. I
suppose I should be happy it works as well as it does.
At any rate, if anyone has any insight here, I'd love to learn more...
- Josh
The fade positive effect is controlled by the OP LEVEL control. I had to
adjust that to get many of my terminals to fully erase and stay that way.
It's touchy. Also, it seems dependent on the room temperature the
terminal is operated in.
Bob
--