-----Oorspronkelijk bericht-----
Van: cctalk-bounces at
classiccmp.org [mailto:cctalk-bounces at
classiccmp.org]
Namens Tony Duell
Verzonden: zaterdag 15 januari 2011 18:20
Aan: cctalk at
classiccmp.org
Onderwerp: Re: HP 8018A heath sink getting hot
I acquired a HP 8018A serial data generator, a very nice piece of
equipment.
But after a little time the power supply heath sink is getting hot
about 70 degrees Celsius (158F).
That does sound rather hot. Personally, if I can't keep my hand on a
heatsink, then I want to investigte (DO NOT try this with SMPSUs where the
heatsinks are often live!)
I'm not sure if that is normal, anyone who can help me with this ?
Ifr you don't already have it, you can get the operating/service manual from
http://www.agilent.com/ . Once you've found the 'manuals page, you enter
the model number (8018) and it'll give you a link to download the pdf
manual. Said manual looks to contain full schematics, etc.
Anyway, I've not really looked at it, but I did notice that the PSU is
entirley linear, and that there's a -5.2V rail, presumably for ECL. To me
that suiggests it might run hot, but how hot I don't lnow.
-tony
I've the manual and did some measurements, there are 3 current sense
resistors one for each supply voltage.
The 25V and -5.2V current are within spec. the 5V current is a little high
but in spec when I remove the HP-IB option card.
ESR of the psu elco's < 0,03 Ohm and tantalium cap's <1 Ohm on the board
are good .
Measuring the heath sing temperature gives 68-70 degrees Celsius, but the
instrument keeps running stable.
So I think it's 'designed' to get hot...
-Rik