ITS WORKING!!!!!!!!!!!!!!!!!!!!!!!!!!!!!
Special thanks to Tony's idea of hooking up an isolation transformer (6.3V
but rated to 3000V)..
Okay, I did some of Christians tests and I found that the Transformer indeed
was arching into the main supply. (I won two of these scopes on EBay).
Here is the funny part.. I rarely go into surplus stores but tonight I made
an exception. I told the guy at the counter what my problem was and he had a
box full of 6.3Volt isolation transformers rated to 3000KV
HAHAHAHHAHAHAHAH.. I could not believe it. I bought a few and tried one out.
The scope now has a perfect trace. On XY the dot is crisp and very stable. I
have to align the scope.. but will do so after I get all brand new tubes in
it.
Second funny part, As I was leaving the store I went over to his lightbulb
section (get the hint yet).. Yes, brand new bulbs [equivs] for my 8/S and my
8-I still in the package!
What a day...
I can't wait to get spacewars up and running.
Thanks for the info Chris and Tony, I have one more scope to restore and
have placed orders for another 8 of them... so I have some serious tube work
over the next couple of weeks.
(comments below)
-----Original Message-----
From: Christian Fandt <cfandt(a)netsync.net
To: Discussion re-collecting of classic computers
<classiccmp(a)u.washington.edu
Date: Saturday,
December 04, 1999 5:34 PM
Subject: Re: Tube experts! - I need your assistance.
Upon the date 10:42 PM 12/3/99 -0500, John B said
something like:
>I don't like tubes. My experience with tubes is *very limited*. I have had
a
few crash
courses over the past few years......
Well John B., I *like* tubes :-) As you may have noticed I'm an electronic
historian and old radio collector. Tubes pervade my very being it seems. I
grew up learning about and fiddling with tube gear :-) Love it! Wish I
could afford getting (or even *finding* an old IBM tube machine like a 704
or such.) A few others here enjoy the same background. Anyway, I went up to
my library and dug out my Tektronix type 503 manual. The RM503 is simply a
repackaged 503 which fits into a 19" rack.
The only reason I don't like tubes is because they are very flakey in old
mini computers.. From what I have heard from people who use to support them
every power cycles was a nightmare. I am trying to stick to minis that can
run off of 110/200Volt. I have a lot of design experience in transistors and
IC/analog stuff... little to no tube knowledge.
Below, I'll speak as if you've had very little
exposure to tube circuits,
'scopes of this vintage, etc. as I really don't know your old technology
background as of yet. At least others here who are not tube savvy and will
someday have to fiddle with an old scope may get something out of this
anyway.
Very limited.. but I have to use these old tube scopes as I am putting the
original tek scopes back into the minis.
>>Okay, I bought the Tek RM503 for my PDP-8/S. When I turned it on I heard
>>some terrible noises and found the power supply voltages were all over the
>>place (way off , like 10V was 500 etc..).
>Not good of course. Limit the on-time while
testing please.
>
I did.. to about 30 seconds... Got lucky though.. the 4 transistors did not
die.
>This scope uses a primary transformer to supply the
6.3V to most of the
>tubes and has a separate winding for 6.3V for the CRT heater. Off this
>transformer another winding fed to a voltage doubler and then to an
>oscillator with another transformer to create a wide range of voltages.
(12V
>>to -3000).
>I sense you have a manual too as you give a
good basic layout of the power
>supply and indicate expected voltages.
Yes, I got it with the scopes.. I do have to buy a bunch of RM560s. Do
youhave any manuals for those?
>
>>Picture this... the -3000 volts is fed
right into the CRT heater (which
>>happens to be directly coupled with the primary transformer). I have been
>>able to locate the problem somewhat. If I remove the -3000 volt line
between
>>the HV rectifier tube and the CRT itself then the scope works fine! All
>>waveforms are proper and the power supply works great (no picture of
>>course).
>So, by removing the -3KV line, things settle
down.
Yes
>First thing in mind is that either of the two pots in the voltage divider
>resistor string may be arcing over to ground. They are the FOCUS and
>INTENSITY controls.
The focus *kind* of worked... nothing on intensity.
(they both work now)
>Second thing and at least this is easy to check, does the CRT heater light
>up? With the -3KV left disconnected from the CRT do you measure 6.3 to 6.5
>volts AC across the heater connections (pins 1 and 14)?
Yes
>Third thing, and most undesireable, is the CRT envelope got broken and the
>tube went to air. This will *definitelly* cause arcing inside the CRT's
>electron gun. You know the shiny metallic spot you often see inside vacuum
>tubes? That's the gettering which basically had taken up most of the
>leftover oxygen after the manufacturer had drawn a vacuum on the tube and
>tipped it off. Never saw a 503 tube so can't say where to find it but it
>usually would be on the inner surface of the neck somewhere maybe 5 or 10
>CM in from the base. If you see a milky white spot on the inner surface of
>the envelope in that region then the tube's gone to air :-(
Thanks.. I wil check out the other scope I have to work on next.
>You mention that instead of 10 volts you measured 500. Thank Heaven this is
>not a solid state scope!!! Smoke City!!
I know.. What scares the hell out of me is I am hooking this thing up to my
8/S.. I am going to put some highvoltage diodes between the 8/S,8I and the
scope to make sure if the scope goes bananas I don't blow a few hundred
transistors in the minis.
>
>>It can be a few things... I am hoping
someone here who use to work on tube
>>units might be able to tell me which problem below it most likely is:
>
>>#1) A bad HV rectifier tube causing the
HV to come back to the second
>>transformer which would put a few thousand volts back into the secondary
>>winding taps causing high voltage everywhere.
>A shorted 5642 HV rectifier tube would
present a high frequency AC voltage
>of some level on the -3KV line. Maybe 5-7 KVAC peak to peak. The freq would
>be 25KHz as generated by the 6DQ6A oscillator tube. Hard to see the little
>filament in the 5642 to verify whether its glowing and you really cannot do
>any measurements with a standard voltmeter on this part of the circuit
The 5642 was glowing and arching inside (looked really bad).. It is happy
now.
(your meter would not tolerate the high voltage unless
you use one designed
for, say, 5KV or more.) I'll assume you may not have such a meter and
cannot measure even the -3KVDC. Check to see if the tube envelope is
I do. I worked in a TV repair shop in my early teens for weekend money.. I
picked up the high voltage meter when the business went south.
>broken. This tube has gettering also and see if it's milky. The filament
>could fail and flop down onto the plate thus making a short circuit. Been
>there, done that. I haven't hunted for 5642's for a long time and they may
>be hard to find now. But I know one of the ham radio community folks could
>come thru if they have a junker Tek 'scope on hand for parts. Let me know.
I found a canadian retailer that has them in stock:
http://www.sphere.bc.ca/test/tekparts.html
also,
http://www.vacuumtubes.com/ has all the tubes in stock (all new in the
box).. It is costing me over $300 for three sets of the tubes though :-(
6DJ8 Amperex tubes are over $30 each :-(
>
>>#2) Bad insulation on the primary
transformer secondary "crt heater"
winding
>>which jumps over to the other winding that happens to be the main
>>powersupply winding (125V X 2)
>Turn off all room lights this evening, close
shades if the city lights are
>bright too. Turn on unit and look for faint arcing around the circuit. Keep
>your hands in your pockets! ;-) You may smell ozone from the arcing. The
>FOCUS and INTENSITY pots are in the resistor divider circuit and are
>mounted on the front panel. Listen carefully to help zero-in on the noise.
>Keep your earlobe in your pocket too! ;-) {ZZZapp!}
>
>>#3) The -3000 volt wire is closely tied to the other low voltage wires. Is
>>insulation breakdown possible due to a crack?
>Yes. Do the lights out trick to verify . .
.
>
>>My next step will be to take a reading on the primary transformer (first
>>transformer, secondary winding [doubled winding]) and see if thousands of
>>volts are there... That might help determine if there is an insulation
>>breakdown but from what I can tell when the -3000V is hooked up every
>>voltage goes crazy.
>
>>I am going to bed.. Hopefully I wake up
to a great answer ;-)
>Hope this helps John. Sorry to be late with
this but family stuff and my
>schoolwork got in the way all day.
Thanks for the info. It is much appreciated. I have a lot of these scopes to
quickly restore....
john
http://www.pdp8.com/
>Let me know if you need other info/help. Regards, Chris
>-- --
>Christian Fandt, Electronic/Electrical Historian
>Jamestown, NY USA cfandt(a)netsync.net
> Member of Antique Wireless Association
> URL:
http://www.antiquewireless.org/