On 2021-03-26 6:08 a.m., Rob Jarratt via cctalk wrote:
I have an old HP 1630G logic analyser. I am
trying to use it to debug
a problem with an 82C206 peripheral controller (or rather I think
damage between the CPU and the peripheral controller). I am not very
experienced with logic analysers and I wonder if I am using it correctly.
What I am trying to do is see which internal registers are being
read/written and the values. To do this there are two signals (XIOR
and
XIOW) that trigger the read/write on their rising edge. So I have
connected the XIOR and XIOW signals to the J and K clock inputs and
set the LA to clock on the rising edge. I have then told the LA to
trigger on a particular address range (in the State Trace screen if
anyone is familiar with this LA).
When I run the analyser it complains of a slow clock. This makes
sense, because I am using the read/write signals to drive the clock
inputs so that I only capture actual reads and writes to the peripheral
controller.
However, I don't seem to be getting sensible
values in the trace and I
am wondering if the LA is really not capturing anything because of the
slow clock?
I don't think it makes sense to clock the LA on the actual clock
signal because I won't be able to capture the address and data values
on the rising edge of the read/write signals and I would end up with
traces full of useless data.
Am I doing it right, or is there a technique that I am missing here?
Thanks
Rob
I think you are the right track, If you wish to only capture register accesses
you may want to qualify on the -ACK signal as the datasheet says it must be
high for register access. I would also clock on the rising edge of the -XIOR
and -XIOW signals as the data sheet seems to indicate that data is not valid
on the falling edge.
Yes, I am using the -ACK signal too as part of the address and clocking on the rising edge
of -XIOR and -XIOW. I wonder if the logic analyser itself might be faulty because I have
set it to trigger on -ACK=1 and XA0-9=08X(hexadecimal, X=don't care), but I get XA
values that do not match the trigger. I am wondering if it is problem with sampling too
late after the rising edge maybe?
Regards
Rob