If anyone with one of those HP1630s with no pods lives reasonably
close to Seattle, we should talk. I have a 1630G that I love (and I
have all the pods) but the screen is starting to dim.... -- Ian
As I am sure you know, the display on the HP1630 is a raster-scanned CRT
monitor. There's conventional-ish video circuitry on the CPU board, using
a 6845 as the iming chain.
The monitor itself seems to have been 'bought in', not designed by HP.
It's the only section for which there isn't a schematic in the service
manual (No, I don't have this on paper, but you can download it fro mthe
Agilent web site [1] :-)).
The monitor looks conventional too. A dism dispaly is most likely to be a
failing CRT, but it could be one of the electrode votlages falling. That
reminds me -- I should measure and record the voltages in my 1630 to use
as a reference if I ever have display problems. When I do this I can post
the results so you can see if yours are at elast reasonable.
I assume you've done the obvious, namely cleaned the screen. Sorry, but
that caught me once...
There are some presets o nthe monitor CPB, and those are documetned in
the service manual. Some instruments have separate brightness presets for
nromal and half-bright text, others have just one control. There's a
procedure for setting them up in the manual which needs a photometer (!),
but it then goes on to say that if you don't have such an instrument,
just do it so it looks good :-). I suspect tweaking these presets might
get a bit more life out of an old CRT.
If you do need to replace the CRT, cmall monochrome tubes are pretty
generic. If it'll fit, has the right deflection angle (90 degrees I
think) and heate voltage (most likely 12V or so) then it'll work. I'd
ratehr take a CRT from a CCTV monitor or something than raid an HP1630
for parts.
[1] YEs, I do prefer manuals on paper, but a web-based manual is much,
much better than no manual at all. Agilent should be aplauded for
supporting their old instruments in this way.
-tony