Ill NLS MS-230
allison
allisonportable at gmail.com
Tue Jul 30 08:22:59 CDT 2019
On 7/29/19 10:30 PM, Jim Brain via cctalk wrote:
> On 7/29/2019 9:24 PM, allison via cctalk wrote:
>>
>> Since you replaced the batteries did you check the fuse?
> Yep, good.
THe batteries are charged? (about 6.6V or more full charge)
>> Also the 4066s
> Hmm, all mine are soldered in. Do you suggest I desolder and check?
>> sometimes get cranky and/or the sockets.
> Only one '00 has a socket.
Check only socketed devices. Do not desolder unless you have good
reason to believe the part is bad.
Hint check for 5-6V on the power pins of the ICs.
IF you have the manual, check for all the voltages.
>>
>> The tough fail would be the CRT. Check the heater pins with a multimeter
>> to see if its open (unplug it as there is a transformer winding across
>> it).
> I'll check.
>> Mine needed one when I got it but they were available then (1977)
>> from NLS. Side effect of the hammer mechanic owner. Now it would be a
>> hunt though old tube brokers.
> Yep, I know that'd be the main issue, but I am hopeful it's something
> less problematic.
The only problem with the CRT is finding a replacement. THey may be
more common than thought. Haven't checked in decades.
Allison